In the manual of the 9227 module from 2010 it is stated a delay time of 38.4/fs+3.2us, on the knowledgeBase article here (Why Is My Data Delayed When Using DSA Devices?) it is 40/fs+3.2us. Which one is the correct one??
06-14-2019 04:20 AM
Hi AndyKr,
did you find the answer ?
06-17-2019 03:25 AM
Hey Charlie,
I'll be surprised if you get a reply on a 7 year old thread, but I'm adding this as a note.
The manual is almost always correct as they are created once and then left, whereas Knowledgebase articles are updated as more information is found that is relevant.
A good example is the article that Andy linked, it was last updated only 2 months ago and no longer references the module that this question relates to. As such, the value from the datasheet should be treated as correct.
Hope this helps!
Mitch
01-12-2022 02:47 AM - edited 01-12-2022 03:05 AM
Hi,
Beware of NI-9227 Datasheet !!!
(last version is dated July 2014).
Why NI used such a stange notation to say "Input delay = 40.01/fs + 2.9 µs" ?
The result is a factor 100 error if you understand the formula as 40 * 5/512 instead of 40 + 5/512.
There is a hudge difference !
Exemple
if fs=2 kHz, you may think that the input delay will be 0.19 ms. In reallity you will observed an input delay of 20.0 ms !
If you have synchronisation considerations in your application, you should be very carefull with this subject.
Hopping that this datasheet will soon be corrected...