Hi There,

 

I am playing with cRIO 9037 & NI 9252 (24 bit AI module).

the problem is that it does not work in FPGA programming mode, please refer to the error message indicated in below picture; but according to cRIO hardware compatibility chart, it should work. https://www.ni.com/en-us/support/documentation/supplemental/08/c-series-module-and-compactdaq-compac...

swangone_0-1605339847461.png

 

swangone_1-1605339924683.png

 

Can you please help to check this problem, or give some hint if there's anything I missed.

 

Regards!

Swangone

0 Kudos
Message 1 of 7
(2,872 Views)

Hey Swangone,

 

The problem might be related to the software version you are using. The 9252 module's FPGA support was added in 2019. You can find more information here: Software Support for CompactRIO, CompactDAQ, Single-Board RIO, R Series, and EtherCAT 

 

 

oscarfonseca_0-1605578490876.png

 

 

Regards,

0 Kudos
Message 2 of 7
(2,823 Views)

Hi Oscar,

 

Thanks for your kind support!

I've reinstalled the software as below, but it still does not work.

 

无标题.png

0 Kudos
Message 3 of 7
(2,815 Views)

I do not have a 9252 with me to test, but I was able to add it manually to LabVIEW 2019. I recommend trying the following:

 

  1. If you right click your FPGA Target and select New > C Series Modules...
  2. Then select New target or device and select C Series Module
  3. Press OK
  4. Select the module type as NI 9252 and Location as Slot 1

 

You should see the following:

oscarfonseca_0-1605657616816.png

 

 

Here are my software versions, in case you want to check if there's any discrepancies that might explain the behavior:

 

oscarfonseca_1-1605657703821.png

 

 

All the best,

0 Kudos
Message 4 of 7
(2,797 Views)

Hi Oscar,

 

I am trying to install the software to be an identical version as yours.

I've installed MAX 20.0 and cRIO driver 20.5 as below, but result in NI-RIO 19.0, I guess this can be the reason that NI9252  is not supported, because a minimum requirement is 19.6., you screenshot also shows NI-RIO 19.6

so, if my assumption is right, how can I get NI-RIO 19.6

swangone_0-1605768074433.png

swangone_1-1605768246638.png

Thanks!

 

 

0 Kudos
Message 5 of 7
(2,774 Views)
Solution
Accepted by topic author swangone

That's very interesting behavior... I installed NI-CompactRIO 19.6 from NI Package Manager. In theory, the NI-RIO driver should be included there, indeed: How Can I Install the NI-RIO Driver? 

 

oscarfonseca_1-1605834979213.png

 

I wonder... did the computer reboot after installing the new driver? If you installed CompactRIO 20.0, I expect NI-RIO 20.0, similar to:

 

oscarfonseca_0-1605834871666.png

 

 

Regards,

0 Kudos
Message 6 of 7
(2,767 Views)

The problem is solved by reinstalling everything to version 20.0

 

Thank you Oscar for your kind support!

Have a nice day!

0 Kudos
Message 7 of 7
(2,746 Views)