09-04-2018 12:07 PM
Thanks. My system is simple. A sensor can output 0-10 V signal, and I want to capture it to NI 9205.
We are seeing the same diagram. Yes, there is a Vcm. I think Vcm is noise voltage instead of a physical device. (Maybe I am incorrect). If you see V2, and its negative is directly connected to AI9, COM, and chassis ground.
I want to connect my sensor only to Al0 and Al8, and leave other channels open, including COM. But I am sure whether this way can make the measurement inaccurate.
09-04-2018 12:25 PM
Some sensors (floating) have two outputs — a + output and a - output. Vcm is not noise exactly. Vcm is common mode voltage—you have to make sure doesn't exceed the spec of 10.4V.
But, it sounds like maybe your sensor has just a + output. If so, that's a single ended source and that's part of what's confusing here of how to wire it. You might consider using the NRSE Voltage Signals wiring diagram
09-06-2018 08:34 PM
Thanks so much! I checked my sensor, it has a positive and a reference ground. Do you think my sensor should be wired using NRSE? Instead of Grounded Differential Signals?
09-07-2018 12:14 PM - edited 09-07-2018 12:15 PM
Your sensor has effectively 3 terminals? Power in, ground, and voltage out? Then, yes, NRSE would be the first thing I tried. I expect it to perform very well. If for some reason it doesn't, the next thing I'd try is Floating Differential Signals