 Ralph@NES
		
			Ralph@NES
		
		
		
		
		
		
		
		
	
			08-01-2007 03:02 PM
			
    
	
		
		
		08-02-2007
	
		
		03:43 PM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 - last edited on 
    
	
		
		
		07-22-2025
	
		
		03:42 PM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 by 
				
		 Content Cleaner
		
			Content Cleaner
		
		
		
		
		
		
		
		
	
			
		
Hi Ralph - 
The 1102 module is made for voltage inputs, so you will have to use that 249 Ohm resistor to measure current.  According to Table 2 of the 1303's Installation Guide, you can leave the 10 MOhm networks in when you connect the sensors.  This will prevent you from having to ground-reference them, too.
08-03-2007 07:10 AM
OK, thanks very much for that info!
One quick question, one of my 1303's has had the resistor networks removed, and sadly lost. I have replacements on the way, but in the meantime what should I do about ground referencing? Connect a 10m from my - terminal to the ground terminal?
Thanks again!
			
    
	
		
		
		08-06-2007
	
		
		05:08 PM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 - last edited on 
    
	
		
		
		07-22-2025
	
		
		03:42 PM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 by 
				
		 Content Cleaner
		
			Content Cleaner
		
		
		
		
		
		
		
		
	
			
		
Hi Ralph,
If you take a look at the 1303's Installation Guide under the following section
Configuring the Resistor Networks » Table 2. Selecting the Appropriate Resistor Networks
you will see that with no resistors in place, you are restricted to using Ground Referenced signals, and with a 10ohm resistor in place you are restricted to using a Floating signal source. I hope that answers your questions.
08-07-2007 07:42 AM
			
    
	
		
		
		08-08-2007
	
		
		11:09 AM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 - last edited on 
    
	
		
		
		07-22-2025
	
		
		03:42 PM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 by 
				
		 Content Cleaner
		
			Content Cleaner
		
		
		
		
		
		
		
		
	
			
		
Hi Ralph,
The following document should clarify the differences between grounded signals and floating signals and how each signal source should be wired.
Field Wiring and Noise Considerations for Analog Signals
Some typical examples of grounded signal sources include thermistors and power supplies.
Some typical examples of floating signal sources include thermocouples and batteries.
Message Edited by Jordan D on 08-08-2007 11:10 AM
 Gubby
		
			Gubby
		
		
		
		
		
		
		
		
	
			07-12-2010 07:47 PM
I too am using a 1303 terminal block with an 1102 module. So, if I have a source that is sending a 4-20 mA signal over a twisted pair, can I simply attach one wire to the negative terminal and one wire to the positive terminal of a channel on the 1303 and that will give me a voltage reading at that channel? I am using the stock resistor networks.
07-13-2010 12:35 PM
Hi Gubby,
Yes, you should be able to wire the twisted pair to the analog input channel, connecting the positive wire to the positive analog input channel and the negative wire to the negative analog input channel. If by "stock" resistor network, you are referring to the factory shipping configuration of 10 Mohms, then you will be configured correctly, whether your signal is floating or ground-referenced.
 Gubby
		
			Gubby
		
		
		
		
		
		
		
		
	
			07-14-2010 05:48 PM
Thanks for the response. Is the current converted to voltage by the pullup and bias resistor networks on the 1303 or is the current passed to the 1102 to be converted there by installing resistors? If the voltage is created by passing 4-20 mA current through 10 Mohm resistors, wouldn't that create voltage that is very high? Also, if this is the case, do the pullup and bias resistors have to be the same (I investigated my setup and it seems that the pullup resistors are 10 Mohm, while the bias resistors are 10 ohm)?
			
    
	
		
		
		07-15-2010
	
		
		05:06 PM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 - last edited on 
    
	
		
		
		07-22-2025
	
		
		03:43 PM
	
	
	
	
	
	
	
	
	
	
	
	
	
	
 by 
				
		 Content Cleaner
		
			Content Cleaner
		
		
		
		
		
		
		
		
	
			
		
Hey,
The SCXI-1102 has a high input impedance. Therefore, it cannot make a current measurement directly. There is a need for external resistors to convert the current to a voltage that can be read by this modules. This is all described in the SCXI-1102 manual in Appendix E.