Signal Conditioning
6-Chan Load Cell Conditioning- Which NI Products?
We are building a machine with 6 strain gage based load cells. We need simple signal conditioning, including power supply, between the 6 load cells and our data acquisition system. What are the NI products we need to buy?
(3,873 Views)
Your might consider the following stsyem:
DAQ Card: 6052E series
Conditioning:
SCXI 1000 Chassis
SCXI-1520 Strain Gauge Module
SCXI-1314 Terminal Block
SCXI-1349 Cable Assembly
This system will give you 8-channels of simultaneous sample-and-hold input with excitation and bridge completion for strain gauges. The 6052E series (PCI or PXI) will give 16-bit accuracy at a maximum rate of 333 kS/sec. Other E-series DAQ cards would also work - it is really a function of accuracy requirements versus cost. As the SCXI bus is only capable of a maximum sample rate of 333 kS/sec, DAQ cards with higher bandwidths would be limited by the SCXI bus.
(3,873 Views)
Thanks for the feedback. However, we don't need the A/D capability. Instead, all we need is to condition the load cells to feed high level (+/- 10V) signals to our existing A/D's. We're trying to keep this really simple and cheap.
Ideas on what product(s) would work?
(3,873 Views)
Sure. For simple signal conditioning, go to www.omega.com (Omega Engineering)and take a look at the following Bridge Conditioning Modules:
DMD-460 Series
BCM-1 Series
OM2 Series
All of these devices provide some/all of the capability you require. Omega Engineering has an extensive line of products to support accurate, economical, and easy pressure, force, and strain measurement. (No I do not work for Omega Engineering; however, I have used their products over the last 20 years and have always been satisfied with the results). Omega will also send a complete reference set of their current catalog in "paper form"(about 70 lbs) upon request. These catalogs have technical instructions and application notes that rival those found at NI
(3,873 Views)
AVR,
Many thanks! Just the info I needed.
(3,873 Views)