Error -200077 scc sg24 with NidaqMx

Why do I get the following error when trying to use test panels in Nidaqmx?

Error -200077 occurred at daqmxcfg_DAQmx Start Task.vi

Possible reason(s):

Requested value is not a supported value for this property.

Property: AI.Max
You Have Requested: 10.0
Valid Values Begin with: -0.1
Valid Values End with: 0.1

Channel Name: SCC1Mod1/ai0

Task Name: _unnamedTask<0>

I am using current verison on Nidaq, 6036E daq card, SC2345 carrier, and SCC sg24 module.
0 Kudos
Message 1 of 2
(2,975 Views)
Hi visual science,

You are attempting to run a test panel on a channel that is connnected to the SG24 strain gague input module on your SCC-2345. The SG24 has an internal voltage gain of 100. This means that the largest voltage you can read through your SG24 is -0.1 to +0.1 volts. After the gain of 100 this will be -10 to +10 volts at the DAQ card.

In the test panel you specified a voltage range of +/- 10 volts. You need to specify +/- 0.1 volts since that is the max that you can read through the SG24.

-Sal
0 Kudos
Message 2 of 2