Signal Conditioning
Loop Ground and Insulation between channel and ground
Hey guys,
I have a problem acquiring the signal from two different accelerometer sensors, where channel 1 is influencing the voltage read from channel 2, either in Differential mode or in RSE mode. I don't know if the problem could be caused by the ground loop or the isolation of the DAQ channels. I will be leaving the diagram of the connections made.
Bruno Lorenzo
(3,439 Views)
What accelerometers and DAQ are you using?
GCentral
(3,428 Views)
ADXL203 and NI6218
(3,406 Views)
Accepted by brunolorenzo
1. You should be using Referenced Single Ended configuration for the DAQ channels.  AI_GND should be tied to the device's ground.
2. Following this circuit note (CN0189), it looks like you should add some amplifiers between the accelerometer and the DAQ.  I didn't dive into the circuit, but you may get away with simple voltage follower op-amp setups.  This was my first thought with your first post as ghosting is common with high output impedance devices.  Adding simple op-amp circuits usually fixes that as op-amps have very low output impedance.
GCentral
(3,402 Views)
Thanks, just adding the voltage follower helped.👍
B. Lorenzo
(3,369 Views)