Hi Will
It is quite difficult to say why you are getting these bad readings. It
could be due to a number of reasons.
Are you using the same Type Thermocouple with both modules?
Have you got your Cold Junction Compensation (CJC) set up?
Have you got any Open Circuit channels on your module?
See
http://digital.ni.com/public.nsf/websearch/91262E2F72A608A9862564F3007B11F6?OpenDocument
Have you set up your second module correctly. I would recommend checking
all your settings in MAX, and then carrying out tests to try and find the problem. See
the following tutorial on setting up the 1102 to make sure it is correctly set
up.
http://zone.ni.com/devzone/conceptd.nsf/webmain/5f569e1eeb62edb68625686600601f68
I hope this helps. Please reply back and let me know how you get on. To reply click the reply link above my name.
Seamus
03-08-2006 10:55 AM
Hi Seamus,
Just to update you on the problem. I found that the resistor bank (pull-up resistors to allow open thermocouple detection) had 10ohm resistors on the low voltage side, whereas on our other 1303 connector block, 10mega-ohm resistors were used. This seemed to be causing the problem. I have read a little bit about this, but I'm still not sure what the difference is, or why the factory default is 10ohm, whereas the reccomended value for thermocouple sensing is 10 mega-ohm.