Signal Conditioning
Ni 9237, c series card
Hello
           iam using Ni9237 card and iam using an 100kg load cell My objective is to measure a weight btw 0 - 50 kg.
iexcitation voltage is 5 v. Sensitivity of the load cell is 2mv/v. reading i get for 0 kg is 0.000646431** v/v
and for 50 kg is 0.00116299** v/v . will i be able measure a weight in grams (or atleast 10 grams ) .
** Places of digit i have mentioned in bold are fluctuating  .
(8,250 Views)
Please confirm the Load Cell specifications
Excitation Voltage = 5 V
Rated Output = 2mV/V
Rated Capacity = 100 Kg
100 Kg Load => 10mV output => 10mV/5V = 2mV/V Output
In your setup, a 50 Kg load instead of 0.00116299 v/V = 1.16299 mV/V
So a 1 g load should give you around 20 nV/V output. Lets see whether it is feasible in NI 9237
As per NI 9237 specifications, the Scaling coefficient = 2.9802 nV/V per LSB (ADC = 24 bits = 224 Levels)
So 20nV/V output lies within the detectable range of NI 9237. You should be able to measure the weight in grams.
(8,237 Views)
Hi Ankur,
              The card has a resolution of 24 bits, but the accuracy is 0.038 mv/v. Is it possible then ?. The value which i have got  fluctuations in the later digits is it because of the accuracy.
(8,233 Views)
Accepted by Rameshs
Yes, the variation in measurement may be primarily due to accuracy.
I think that load cell specifications contain accuracy details. Generally it is in % of maximum rated weight.
Trust that helps
(8,230 Views)
Good day
I am to write a code that will read data from two NI 9237 modules (6 channels in total). the modules are to read data from a force balance (strain due to force).
I know its not a matlab community but I could get help, how do I convert strain result to force using a loadcell of specification stated below and the NI 9237. being specific, I need the coefficient to convert strain value given by the daq to force.
Load Cell: STA-7 Aluminum Alloy S Type Low Range Tension/Compression Load Cell
Rated load (kg)                      1, 2, 5, 10, 25, 50
Ultimate breaking load         300% of rated load
Maximum transverse load   <50% of rated load
Output/ Sensitivity                2mV/V (±0.1%)
Zero balance              <±1% of rated load
Excitation voltage                  10vdc recommended, 15vdc maximum
ch6.BridgeMode = 'Full';
ch6.ExcitationVoltage = 10;
ch6.NominalBridgeResistance = 350;
(3,649 Views)
