Hi,
I have a question regarding the SCC-RTD01 temperature module. I connected a RTD to the module, ran the LabView example, and set the excitation current (external) to the recommended 1mA. I was happy to see a value of 25 deg C. I figured since I hadn't really calibrated anything being off by 2 degrees was no big deal. At one point I decided to try reducing the excitation current to 0.9mA and got a temperature reading of 56 deg C. Is this normal for the excitation current to change the measurement so drastically?
Thanks,
Mark
08-10-2005 09:59 AM
08-12-2005 05:29 AM
Yes that is normal
As already recommended, try to use the internal current source, than you can expect the given spec.
U=R*I and the R of your RTD is only small but well known changing with temperature.