Signal Conditioning
Reading a thermocouple with the SCXI 1520
Does the SCXI 1520 have something to account for the interface connection heat, and is it needed? I've heard this is what the thermocouple modules do, and it is necessary.
(3,247 Views)
The SCXI-1520 uses the SCXI-1314 Module which does not have cold-junction(CJ)compensation capability - meaning it can not compensate for either the temperature of the connection to the DAQ system, nor, the additional voltage effect due to the dis-similar metal contacts of the DAQ system. This correction is absolutly necessary for accurate temperature measurement. I suggest you use a "third-party" cold junction module from a vendor such as OMEGA Engineering (www.omega.com). The CJ sensors are battery powered and reference the signal to 32 Deg F while at the same time providing standard copper leads for the DAQ system.