SCXI-1102 virtual channel reads OK but SCXI temp scan.vi does not

I am using a SCXI-1102, with a SCXI-1303 terminal block. I am not getting an accurate reading using SCXI Temperature Scan.vi. I am trying to use this VI with the channel string ob0!sc1!md2!0. As a test I configured a virtual channel 1102_0, for Deg F, min 0, max 100, J-type, and CJ Source Built-In. I am getting an accuarate reading with this channel. I will attach a VI showing this. I saved the last readings I got as the defaults so that the differences can be viewed. Can I get an advisement on what I need to do here. I do not want to use virtual channels because I want to give the use the choice of temp units, etc. Also, I have tried using the example program that ships, which is for u
sing an 1102 and reading thermocouples. This example is giving the same results as when I use SCXI Temperature Scan.vi.
I have called about this several times on another support request 7011362, and I have spent at least a full day on this problem, I need to get a solution, help is greatly appreciated. I will also attach a screen shot of the config on my virtual channel.
Download All
0 Kudos
Message 1 of 2
(3,084 Views)
Hello Drew,

Use the SCXI-1102 Thermocouple.vi which can be found in the examples (Help>> Find Examples>> Hardware Input and Output>> DAQ>> SCXI)

Make sure the J Type thermocouple is selected and the CJC type is set to thermistor and not IC sensor. This will give you readings in Degrees Celsius.

Regards,

Gerardo Garcia
National Instruments
Applications Engineering