Signal Conditioning
SCXI-1121 - RTD 3 deg F diff between cards
Background:
I have installed 2 new SCXI-1121 cards in a SCXI-1001 chassis.  All channels are configured to measure temp using RTDs.  The cards are configured identically (Gain 10x/10x, filter 4 Hz/4Hz, Excitation current mode = 0.15mA).  Using MAX or LabView, the temp readings differ by 3 deg F at ambient (70 deg F) between the two cards.  I have placed a precision 100 ohm resistor on the TBX-1328 terminal block and connected it to the 1121 and confirmed that the temp difference between cards is 2+ deg F.  One card reads 31.7 deg F, while the other reads 29.2 deg F.  All 4 channels have about the same offset.  I have swapped the cards in the chassis and the temp diff moves with the card.
Questions:
Do I have a calibration issue?  Could I have a conflict between the physical jumpers on the card and setup in MAX?  What is the recommended gain/filter/current for this application? The calibration procedure addresses each channel separately.  I have checked the excitation current and it appears to be correct at 0.15mA using a handheld meter.  It reads between 0.14 and 0.15 mA.  In order to perform a calibration, it appears that the 1121 card needs to be extended in the chassis.  I don't find a SCXI card extender on the web.  Do you have a part number for this item?
(5,114 Views)
Hi Wantacobra,
Great job with the debugging! Now that we know the issue follows one 1121 module, does this difference happen on all channels of the 1121? If not, than I would suggest switching around RTD’s as you may have a bad one. If it does happen on all channels, calibration is a good idea. Please see this knowledgebase: Compensating for the Offset on an SCXI-1121. It has links for self calibration and Calibration Solutions from National Instruments.
Message Edited by David L. on 08-24-2007 12:12 PM
David L.
Systems Engineering
(5,109 Views)
Yes all four channels have about the same offset.  I just found the article on compensating for the offset on an SCXI-1121 and will be trying the second method, using the EEPROM on the 1121.  Thanks for the prompt response and help David.
(5,101 Views)