Background:
I have installed 2 new SCXI-1121 cards in a SCXI-1001 chassis. All channels are configured to measure temp using RTDs. The cards are configured identically (Gain 10x/10x, filter 4 Hz/4Hz, Excitation current mode = 0.15mA). Using MAX or LabView, the temp readings differ by 3 deg F at ambient (70 deg F) between the two cards. I have placed a precision 100 ohm resistor on the TBX-1328 terminal block and connected it to the 1121 and confirmed that the temp difference between cards is 2+ deg F. One card reads 31.7 deg F, while the other reads 29.2 deg F. All 4 channels have about the same offset. I have swapped the cards in the chassis and the temp diff moves with the card.
Questions:
Do I have a calibration issue? Could I have a conflict between the physical jumpers on the card and setup in MAX? What is the recommended gain/filter/current for this application? The calibration procedure addresses each channel separately. I have checked the excitation current and it appears to be correct at 0.15mA using a handheld meter. It reads between 0.14 and 0.15 mA. In order to perform a calibration, it appears that the 1121 card needs to be extended in the chassis. I don't find a SCXI card extender on the web. Do you have a part number for this item?
08-24-2007 12:11 PM - edited 08-24-2007 12:11 PM
Hi Wantacobra,
Great job with the debugging! Now that we know the issue follows one 1121 module, does this difference happen on all channels of the 1121? If not, than I would suggest switching around RTD’s as you may have a bad one. If it does happen on all channels, calibration is a good idea. Please see this knowledgebase: Compensating for the Offset on an SCXI-1121. It has links for self calibration and Calibration Solutions from National Instruments.
Message Edited by David L. on 08-24-2007 12:12 PM
08-24-2007 03:19 PM