Signal Conditioning
SCXI 1125 self calibration
Does SCXI  1125 support self calibration? If yes, can I do it using the DAQmx Self Calibrate.vi?
"A VI inside a Class is worth hundreds in the bush"
യവന്‍ പുലിയാണു കേട്ടാ!!!
(6,588 Views)
Following is the  link that'll help you out.
http://zone.ni.com/devzone/cda/tut/p/id/4845
However, whenever you use the 11254 with the SCXI-1327, you should be careful about the following.
Whenever you try to perform a Self-Calibration on a SCXI-1125 module with a SCXI-1327 accessory, it will only pass if you have identical attenuation settings on every channel. As soon as you change one of the channel attenuation to be different from the other channels, the self-calibration will fail.
Hope this helps.
Vishal
(6,577 Views)
Thank you Vishal.
SCXI 1125 user manual says that internal calibration is possible (for offset) and that we should use the SCXI calibration VI (not the DAQmx Self calibration VI). I am not able to find this VI in the DAQmx Calibration palette. The only SCXI calibration VI present is for external calibration.
(6,570 Views)