03-23-2009 02:24 PM
HI ,
I ned to measure lod from a 1000 ohm full bridge circuit. The distance between the sensor & my SCXI 1520 is 500 meters. Please let me know how can I compensate for the lead resistances both in Excitation & while reading back the actual sensor value.
03-24-2009 12:13 PM
grte,
For extremely long cable runs resulting in much higher than normal lead resistance to bridge resistance ratios, you must use remote sense, which is a six wire connection (four wires for excitation and two for signal). Refer to your SCXI-1520 manual for more information: http://www.ni.com/pdf/manuals/372583d.pdf, specifically page 2-9 and Chapter 4, Theory of Operation.
03-24-2009 02:29 PM
Hi,
Thanks for your response. It would be of great help if you can show me a practical example. My leadresistance for a 100 meter cable will be around 8 ohms. PLease let me know the maximum cable length the SCXI 1520 can support.