Sampling multiple SCXI analog modules

I need to sample 32 channels from 4 SCXI-1125 modules at 50Ks/s per channel.I'm thinking of the following configuration:

4 SCXI-1125's in parallel mode
2 PXI-6259's with 4 cables to the SCXI boards (each 6259 has 2 connectors).

This should give me max sampling rate of 1MsS/16 = 62.5Ks/S.

Will this configuration work? (there is a rule 1 SCXI per 1 DAQ device in parallel more, but is seems that the 6259 board can connect to 2 SCXI)
0 Kudos
Message 1 of 2
(3,043 Views)
Daniel,
That setup should work fine. And you are correct that the one DAQ device per one SCXI module in parallel rule does not apply to M Series devices with two connectors. One last thing I'd like to point out is that when you are determining your sample rate, you do not need to divide the total sample rate by 16, but only 8. Even when you are in differential mode (which you will be in this setup), which means of course that you are using two analog input channels for each measurement, it does not add the overhead of sampling two channels. So your max sampling rate would be more like 1MsS/8, or about 125kS/sec. Hope this info helps!
-Alan A.
0 Kudos