Hi wag,
J9 is used for digital I/O or second stage AI, that is the likely reason why you're obtaining an error. You should use J2 for your second module instead. Since the TC-02 module uses two channels (for TC measurement and CJC), hence when you use 8 of these modules, you're using 16 AI channels of your DAQ device. If you need to use 10 TC-02 modules simultaneously, the M Series devices with 32 channels will allow you to do so with only one board. Please see the article below:
Using 32 Channels Of An M-Series Card With SC-2345The guide will explain the signal routing of different slots:
SCC Quick Start Guide Thanks,
Lesley Y.