Signal Conditioning
Some trouble with my SCXI 1520 module Using MAX4
Hi !
I connected a full bridge Strain gauge   to the SCxi 1520 module but  now i got some problems  .
1-do I have to set the gain of the amplifier first( can't see how)and then create the scale  or it is not necessary
2-how can I calculate the shunt resistance( I am using 4 resistances of 350 ohm for the full bridge), do i have to do this or it is done automatically .
I will appreciate your help
Thank u
(3,613 Views)
Problems 2:
Each input channel of the SCXI-1520 includes two independent shunt calibration circuits, with which you can simulate two separate loading effects on your strain-based device and compensate for any possible gain errors.
The RN-55 style shunt calibration resistors are in sockets
and located in the SCXI-1314 front-mounting terminal block.
You enable or disable the shunt resistors through software commands.
(3,589 Views)
Looks like u have not understood the funda of shunt cal
read this link to understand about strain measurement and shunt cal
https://www.ni.com/en/shop/data-acquisition/sensor-fundamentals/measuring-strain-with-strain-gages.h...
(3,581 Views)