06-01-2006 08:33 AM
06-02-2006 10:13 AM
Surender,
Have you performed a strain calibration on all of your channels. Also, you should be sure that your connections and resistance values match those specified in the document linked below. Please verify this and let me know if you are still having problems.
http://www.ni.com/pdf/manuals/372583d.pdf
Regards,
lrallen
06-03-2006 12:50 AM
Thanks Irallen,
As for channels calibration issue, i have check my all channels but found same behaviour.
I am sure about connections and my straing gauges are all of 120 ohms.
I have problem still also.
Surender
06-05-2006 04:17 PM
Surender,
Take a look at this KB, it looks very similar to the issue that you are having. -GDE06-06-2006 11:38 AM
06-14-2006 02:50 AM
Thanks Logan,
I am agree with you but if i need to take reading in the same fashion then what should i do?
Shall i use wheatstone bridge equation in place of ni supported 7 user modes q-i, q-ii, half-i etc. if yes then how?
Surender
06-15-2006 12:59 PM
Surender,
I am a bit confused by your last post and what you are
asking. I believe
06-16-2006 01:06 AM
06-16-2006 03:30 PM
Surender,
I think I may see the confusion. When you choose strain/bridge type then the equations are taken care of for you. That is what the DAQmx driver is doing behind the scenes. So choose what kind of reading you want to make, and the driver will apply the appropriate equations, and give you back a strain measurement. Take a look at this tutorial, it may shed some more light on strain measurements in general.
-GDE