Hi, we used PXI-4462 for data acquisition when measuring sound pressure. It worked good.
But for make our device more portable, we bought USB-4432, and want to use it instead of PXI-4462. The problem now is that the data obtained through USB-4432 have high level background noise, which didn't happen before.
The thing is, we used the same program for PXI-4462 and USB-4432, the data from PXI are good, but Data from USB have high level noise. Is there any special set needed for USB-4432? or the data acquisition qualities are different between PXI and USB?
07-14-2009 02:37 PM
07-28-2010 09:42 AM
The voltage range on the USB-4432 is +/- 40 V which is probably too high for the sensitivity of the microphones that you are using which ranges up to 50 mV/Pa which means the signal is typically below 5 V at up to 133 dBA. The USB-4431 or USB-9234 are better choices, but the NI 446X and NI 449X series are our best choices with multiple gain settings and higher dynamic range.