Signal Conditioning
Use of different sensors with scc-sg24
in my application I've to use two type of sensors with following characteristics:
-a pressure sensor : 10V Ex V , input nominal 950 Ohm output nominal 300 Ohm
(what does it mean? I know that should be a 350ohm full bridge...)
-a load cel: 10V Ex V, 700Ohm full bridge
It is the SG24 able to read this kind of sensors?
Can I use a single SG24 module to read the two sensor using the two
channels at same time?
Thank you,
Pets
(7,225 Views)
Hi Pets,
by looking at the scc-sg 24 specifications at this link https://www.ni.com/docs/en-US/bundle/371073b/resource/371073b.pdf
you can see that your module has a 10 V excitation which is the same amount required by your sensor.
The module will work with either the 350 Ohm resistance and 700 Ohm, but since it's been designed for a 350 Ohm bridge, this will be more sensitive.
I hope this informations will help you.
Regards
Andrea N.
Principal Applications Engineer - Semiconductor EMEA
National Instruments Italy
Certified LabVIEW Architect - Certified TestStand Architect
(7,218 Views)
Hi Ryan,
the sensitivity leak is not a problem.
But it is not clear if I can connect the two sensors at same time.
Thank you,
Pets
(7,215 Views)
Accepted by Pets65
Hi Pets,
the SCC-SG 24 has two differential channels so you can connect the two sensors at the same time.
Ciao,
Andrea N.
(7,213 Views)