Hi,

in my application I've to use two type of sensors with following characteristics:

 

-a pressure sensor : 10V Ex V , input nominal 950 Ohm output nominal 300 Ohm

(what does it mean? I know that should be a 350ohm full bridge...)

 

-a load cel: 10V Ex V, 700Ohm full bridge

 

It is the SG24 able to read this kind of sensors?

 

Can I use a single SG24 module to read the two sensor using the two

channels at same time?

 

Thank you,

Pets

 

0 Kudos
Message 1 of 4
(6,954 Views)

Hi Pets,

 

by looking at the scc-sg 24 specifications at this link http://www.ni.com/pdf/manuals/371073b.pdf you

you can see that your module has a 10 V excitation which is the same amount required by your sensor.

 

The module will work with either the 350 Ohm resistance and 700 Ohm, but since it's been designed for a 350 Ohm bridge, this will be more sensitive.

 

I hope this informations will help you.

 

Regards 

Andrea N.
Principal Applications Engineer - Semiconductor EMEA
National Instruments Italy
Certified LabVIEW Architect - Certified TestStand Architect
0 Kudos
Message 2 of 4
(6,947 Views)

Hi Ryan,

 

the sensitivity leak is not a problem.

But it is not clear if I can connect the two sensors at same time.

 

Thank you,

Pets

0 Kudos
Message 3 of 4
(6,944 Views)
Solution
Accepted by topic author Pets65

Hi Pets,

 

the SCC-SG 24 has two differential channels so you can connect the two sensors at the same time.

 

Ciao, 

Andrea N.
Principal Applications Engineer - Semiconductor EMEA
National Instruments Italy
Certified LabVIEW Architect - Certified TestStand Architect
Message 4 of 4