Signal Conditioning
We have difficulties with using SCXI-1000, two SCXI 1102 and two SCXI-1303 with one 1349 and a DAQCard-6036E
We want to use the two 1102 and 1303 with one DAQCard. And now we do not know how to configure that.
1. Is there a connection between the two 1102C's and the 1349 required? Every 1102 works fine when we are testing one after the other with a SCXI-1349 in the back (MAX) parallel. -> Test Panel. But we want to use both 1102 with one 1349.
Thank you very much in advance, Andy
(3,656 Views)
Hi!
You have to connect the 1349 to the the scxi card which is on the left side of the chassis. In Max you have to injust multiplexing acquisition (not parallel!)
If you have more questions you should read the "Getting Started with SCXI" manual
regards,
Thomas Sandrisser
NI Germany
(3,656 Views)
The connection wire between the two 1102C's and the 1349, is it included in the usual delivery package of 1349?
(3,656 Views)
Why you want to connect the 1102cards with a cable?
If you have your scxi1000 chassis in front of you, put the cards into the slots! these cards are interconnect by the scxi bus!
For the communication with the daq card you need the 1349block. you have to put this block on ONE of the scxi cards (preferred to this one, which is on the left side!)
In MAX you have to configure which card has got the 1349block...
After that it's possible to communicate with all scxi cards in your chassis.
regards,
Thomas Sandrisser
(3,656 Views)