08-08-2005 04:48 PM
I need to perform a shunt calibration on an SCC 2345 SG-24 strain gauge module with a load cell transducer. Apparently shunt resistors are not included with the modules so I need to buy some. My load cell has 350ohm bridge legs. I was previously told that precision resistors need to be used for strain gauge applications. What accuracy is considered precision, 1%, 0.5%, 0.1%? What diameter lead wires are needed?
I was unable to find anything on the NI website. Can anybody recommend a supplier?
08-09-2005 09:11 AM
08-09-2005 09:21 AM
08-09-2005 01:42 PM
Sorry, I guess I was too quick with my message and didn't double check it. I actually need to do an offset null adjustment (which the SG-24 doesn't include resistors) with a 30k ohm resistor and then do a shunt calibration at 60k ohms. The SG-11 has a 301k ohm resistor.
So knowing that I need a long term stability and low temperature coeffienent, what % accuracy resistance I would need for the offset nulling? I see from the manual that the shunt resistor is 1%.
08-09-2005 01:44 PM
08-10-2005 04:13 PM