I may not have been clear in the previous post.
I already tried disconnecting the original SCXI-1300 terminal block (with all of the RTD and other instruments connected) from the old SCXI-1102C module, then connected a second/new SCXI-1300 terminal block. I then tried shorting and sourcing various channels in the new terminal block and read the corresponding voltages in the M & A Explorer as well as the machine's VB calibration and diagnostic application. This should have eliminated any errors associated with a grounded signal, unexpected open channel, defective terminal block, etc. The same offset still occurred.
Going a step further, I replaced the old SCXI-1102C with a new unit, still using the new SCXI-1300 terminal block with shorted and sourced channels, and still measured the same (-100mv) offset. This suggests that the original 1102 module is not the source of the problem.
Next, to eliminate the possibility that errors were originating from other modules connected to the SCXI-1001 chassis (slots 2 through 6) all of the other modules were unplugged to note any impact on the voltages read in slot 1; no difference.
Hopefully this is more clear. I'm thinking that the problem is in the SCXI-1001 chassis (power supply, timing circuitry or multiplexer) or the PCI-6052E DAQ device. Anyone agree? Other ideas?
Regards,
Jeff