Hello,
Thank you for contacting National Instruments.
The first thing that should be noted is that the SCXI-1308 terminal block is used for current inputs. The SCXI-1308 has 249Ohm shunt resistors through which your input current is passed to produce a voltage. This voltage is then measured by the SCXI-1100. This means that you should be able to measure your current signals, but in order to effectively measure your voltages, you will need to remove the shunt resistors for each voltage input channel you have connected to your SCXI-1308.
The next thing that should be noted is that the SCXI-1100 is a jumper programmable module. This means you have to set jumpers inside the SCXI-1100 module and then make sure that the jumper settings match the configuration settings in MAX. Here is a link to the SCXI-1100 User Manual if you need information on how to set the jumper properly:
http://digital.ni.com/manuals.nsf/websearch/0790D3C92CF4B16E8625665E00712865?OpenDocument&node=132100_US
If you are using the SCXI channel string to address individual channels, you will need to make sure you use the following string: scX:mdY:Z where X is the chassis number (ie. 1), Y (ie. 2, 5, 8, 11) is the module number, and Z is the channel number (ie. 0, 11, 19, 31)
Also, if you have other modules in your system, you will want to make sure that you have your PCI-6023 cabled to the correct module. Please see page 10 of the SCXI Quick Start Guide to ensure that you are connected to the correct master module:
http://digital.ni.com/manuals.nsf/websearch/7974503CC2946D0886256E6300748710?OpenDocument&node=132100_US
I recommend that you that you delete your chassis configuration from MAX, verify the jumper settings in each SCXI-1100, make sure that you are cabled to the correct master module in the SCXI-1001 chassis and also make sure that you have removed the 249Ohm shunt resistors from the channels of the SCXI-1308 to which you have voltage signals connected.
Regards,
Bill B
Applications Engineer
National Instruments