10-17-2006 11:31 PM
10-20-2006 06:06 AM
Hi,
The input resistance will be related to the SCXI-1125 input module and not the SCXI-1313 attenuating module.
The specifications for the 1125 can be found here - http://www.ni.com/pdf/manuals/322425a.pdf, with the information you are looking for being on page 47.
The impedance is listed as :
Powered on ..................................... 1 Gohms
Powered off..................................... 4.5 Mohms
With SCXI-1327 ............................. 1 Mohms
Hope this answers your question 🙂
Tom.
NIUK
10-20-2006 02:25 PM