Signal Conditioning
input impedance with SCXI-1313
I am using SCXI-1125 with SCXI-1313 as the terminal module.
What is the input impedance when 1313 is set to attenuation mode (0.01 gain) and no att. mode (gain = 1)
Rg
naushica
(3,737 Views)
The input resistance will be related to the SCXI-1125 input module and not the SCXI-1313 attenuating module.
The specifications for the 1125 can be found here - http://www.ni.com/pdf/manuals/322425a.pdf, with the information you are looking for being on page 47.
The impedance is listed as :
Powered on ..................................... 1 Gohms
Powered off..................................... 4.5 Mohms
With SCXI-1327 ............................. 1 Mohms
Hope this answers your question 🙂
Tom.
NIUK
(3,720 Views)
I appologize for contradicting the previous answer, but when the SCXI-1313 is using the 100:1 attenuator, it uses a voltage divider which will result in an input impedence of 1Mohm. When the attenuation is 1:1, the input impedence will equal the impedence of the SCXI-1125 (~1Gohm).
The SCXI-1313 installation guide describes this attenuator resistance:
http://www.ni.com/pdf/manuals/322853a.pdf
-Logan K.
(3,711 Views)