12-17-2009 08:31 AM
Hi.
I need to measure the wall crack with LVDT but the measurment distance too long. That is aproximitly 70 m.
If i use SCXI-1540 due to voltage drop what it will happen.
12-18-2009 03:55 PM
12-28-2009 03:18 AM
So how long can the maximum SCXI to DAQ card cable use to connect these device each other.
In this case i m using two SCXI chassis, some input module and PCI/ USB- 6281 daq card.
I found one cable for connecting that is SCXI-1353 Shielded Cable, 10 m
but this distance so short for me maybe there is longer cable from that as customize
12-29-2009 07:40 AM
Hey Teknosat,
Take a look at the KnowledgeBase below for more information on this.
Thanks again,
Nick
01-07-2010 04:28 AM
How about using a signal conditioner near the sensor that output a nicely scaled current loop?
If you are stuck with that equipment: A lot of (digital ) filtering and a recalibration of the setup (appropriate fixure and two gauge blocks) . However with 70m (hopefully known type of) cable at not stable temperature I would also estinate the voltage drop of your exitation over temperatur ..