Signal Conditioning
output voltage drop in USB-6212
@LindsW wrote:
Hi Jonnas,
It sounds like a unity gain op-amp (voltage follower) is the correct approach.  Take a look at this white paper article, which discusses how to hook up a DAQ device with a voltage follower (section 3) to double check you have it connected correctly:
https://www.ni.com/en/support/documentation/supplemental/12/using-a-unity-gain-buffer--voltage-follo...
Thanks Linds!
AS a matter of fact, I did go the the op-amp road but it turned out that the
follower op-amp indeed follows the voltage drop... Smiley Surprised
I figured that the real problem I have is the 2mA limit on the USB-6212
So I designed the simplest emitter follower transistor current amplifier [examle in the link]
(http://www.colorado.edu/physics/phys3330/PDF/Experiment7.pdf)
and for now it works perfectly.
I'm happySmiley Very Happy
Thanks for a nice discussion!
-J
Message 11 of 11