10-07-2005
04:08 PM
- last edited on
11-27-2025
06:58 PM
by
Content Cleaner
10-10-2005 08:42 AM - edited 10-10-2005 08:42 AM
Message Edited by Logan K on 10-10-2005 08:53 AM
10-10-2005
09:46 AM
- last edited on
11-27-2025
07:04 PM
by
Content Cleaner
Logan,
Thanks for your response.
At this point, the software is written with Traditional NI-DAQ. I am looking into having it updated to use NI-DAQmx, but that's not a guarantee at this point. Another thing that I failed to mention in my previous post is that I use this acquisition system at different sampling rates, with 10kHz being the max sampling rate that I'm looking for. Given that, I want to have programmable anti-aliasing. That's one of my main reasons for going to SCXI. (I currently have a 6071E DAQ board with a 5B backplane for signal conditioning.)
So it looks like I have no choice but to cascade SCXI modules to get the functionality that I desire. What types of problems will I run into with this type of configuration? From the article here: https://www.ni.com/en/support/documentation/supplemental/06/cascading-scxi-modules.html it seemed like a pretty straight-forward solution. I believe that with our Acquisition software, we still use Measurement and Automation explorer to configure the hardware, and then the program simply interfaces with the DAQ board (i.e., E-series device).
In addition, this article: http://digital.ni.com/public.nsf/websearch/1936a86d5321444186256ff1004ff966 seems to indicate that both connectors on the M-series board can be used with SCXI modules in parallel.
At this point, I'm not sure what my best course of action is. From the articles above, I was under the impression that there was a pretty straight-forward solution to my needs, provided I could get our software upgraded to NI-DAQmx, but now it appears that may not be the case. Do you have any other advice on how I should proceed?
Andrew Krajnik
10-11-2005 09:14 AM
10-11-2005 12:11 PM
Logan,
Thanks again for your prompt and comprehensive response. The further I have investigated this solution, the more conflicting information I've been getting, and I appreciate you taking the time to bring some clarity to the situation.
As far as the 4472 goes, I like the fact that it is a concise solution, but as you stated, it is overkill for my application, and I don't think I'd be able to justify the added cost.
Let me try to summarize my understanding at this point:
1) If I want to use SCXI to achieve simultaneous sampling, I would need to use 1 M-series board per SCXI-1140, since the second connector on the M-series card doesn't support it.
2) If I want to cascade from SCXI-1120, to 1141/2/3, to 1140, my accuracy will suffer due to the fact that I am not reading directly from the 1141/2/3, and don't have the onboard constants available.
*** A question about "accuracy": when you say my accuracy will suffer, you're referring only to the absolute voltage measurement, right? If I don't care about exact voltage, and I calibrate the entire system based on my Engineering units, will I see any difference? Or is the linearity affected as well?
Are there other problems with cascading, if I can live with the accuracy issues you mentioned? If I were to simply cascade the modules in the SCXI chassis, and program the 1141/2/3 filters using MAX, would I simply be able to treat the setup as though I were only acquiring from the 1140?
I guess at this point, I see 2 options that could get me where I need to be:
Option 1: Assuming that I can tolerate the loss of accuracy, and assuming that cascading gives me no other programming problems, for each 8 channels I could cascade 3 SCXI modules to 1 M-series board, and for 32 channels, I'd have a total of 12 SCXI modules, and 4 DAQ boards.
And the other option, that just occurred to me:
Option 2: I could cascade SCXI-1120D to 1141/2/3 modules to an S-series board. For 32 channels, I'd have 8 SCXI modules, and 4 DAQ boards. Again, correct me if I'm wrong, but by doing this, I'd be acquiring directly from the filter modules, so my accuracy wouldn't suffer. Do you see any problems with this configuration? If not, it seems like a pretty attractive option.
Am I correct in my logic here, or is there something else that I'm missing?
Andrew Krajnik
10-14-2005 04:35 PM
10-16-2005 09:34 PM