I have system
1- 65 analog input
2- 15 analog output
3- 20 digital i/o
If I use SCXI 1600 data acquisition I need another analog input module or analog output module or digital module
(4,197 Views)
65 analog input:
You could use 2 32 channel modules (SCXI-1100, 1102/B/C, 1104/C).
15 analog output:
Would require 3 6 channel modules (SCXI-1124).
20 Digital I/O:
Would require either a SCXI-1162 (DI), SCXI-1163 (DO), or a combination of the two.
Worst case, this is 7 modules that you need to control. You would need at least 8 slots including the SCXI-1600 controller. The SCXI-1001 is a 12 slot chassis that would work for this setup.
Does this answer your question? If not, please reply with more details and I will be glad to help you out.
Sean C.
(4,193 Views)
I reduce the specifications of our system to
1- 55 analog input
2- 5 analog output
3- 18 digital output
Do we need SCXI in this case or not?
What is good solution?
(4,184 Views)
Aisle sorry i change Alias to engmoh
(4,181 Views)
(4,181 Views)
In this case,
55 analog input:
You would still need 2 32 channel modules (SCXI-1100, 1102/B/C, 1104/C).
5 analog output:
Would only require 1 6 channel module (SCXI-1124).
18 Digital I/O:
Would still require either a SCXI-1162 (DI), SCXI-1163 (DO), or a combination of the two.
Worst case, this is 5 modules that you need to control. You would need at least 6 slots including the SCXI-1600 controller. The SCXI-1001 is a 12 slot chassis that would work for this setup.
Keep in mind, that if you do not need any signal conditioning, you can achieve this channel count with 2 PCI/PXI-6259s. They have 32 AI, 48 DIO, and 4 AO channels each. For more information on this product please click on the following Link.