Rajeevan,
Sorry this issue has been so frustrating. Before I attempt to answer your question let me see if I understand what is going on. You have connected your load cell up to a 1314 that is connected to the front of the SCXI-1520. The SCXI-1520 is located in the left-most slot of the SCXI portion of your PXI-1052 chassis. The SCXI-1520 is controlled by the DAQ card in the right most slot of the PXI section of the chassis. You have created a task in Measurement and Automation using the following parameters:
Type of Task: Custom voltage with excitation
Range: 200 to -200 kg
Prescaled Range: 0.002 to -0.002
Excitation: 2V
When you read the channel it returns 200. This means the channel is railed, or in other words the voltage present at the channel is greater than the max prescaled value or 0.002. The driver will clip any value that exceeds the specified limit. If you use a battery as you have mentioned the voltage will far exceed the 0.002 V limit. You cannot use your task to test the battery. You should just use the test panel. You can get the test panel by expanding the Devices and Interfaces folder under My System. Then expand the NI-DAQmx Devices folder. You should see the SCXI portion of your PXI chassis listed. Expand the chassis to find the 1520 in slot 1. Right click the 1520 module and select Test Panels... Specify a Max and min voltage greater than the voltage of the battery. Apply the battery across S+ and S- with the load cell disconnected.
The DAQ diagnostic utility is actually very useful and simple way to diagnose possible problems. It is simply an executable that you run on your system that automatically checks your card. You do not need to connect external signals or do anything complicated. Just run the program. You can get the DAQ diagnostic utility at:
http://digital.ni.com/softlib.nsf/websearch/C8D5A739965EF222862571D900780D3CDownload the zip file and run the diagnostic utility. Make sure nothing is connected to the front of the DAQ card before running it. The utility will generate a report and will tell you if the card is functioning properly.
What is the specific load cell you are using? Could you give me the manufacturer and module so I can take a look at the specs? Different load cells have different excitation requirements and this could also be a potential problem. I also want to confirm that the correct range and scale were specified.
How did you specifically wire the load cell? Which wires did you attach to each pin on the 1314 connector block?
You also made mention that the multimeter read zero volts when nothing was applied and also for when 10 kg was applied. What range did you have selected on your multimeter. The top of your range is only 2 mV and will register as 0 V if you set your DMM to a large range.
Let me know if I misunderstood some part of your application. Also let me know if you have questions about the things that I mentioned. Good luck with your application.
Regards,
Neil S.
Applications Engineer
National Instruments