wrong readings from a thermocouple

hello
Im getting wrong readings from a thermocouple T type, installed on SCXI1100/1300 using MIO6070E
Tareq
0 Kudos
Message 1 of 2
(2,979 Views)
Tareq,

first test: shortcut your TC input . Can you read the temperature of your input terminal (resp. of the tempsensor for the cold junction compensation)?

If no then you have a problem with your measurement setup.
If yes then the problem is you sensor (signal)
maybe to much noise due to ground loops or other EMC
do you use a extension cable? Right type ? Right polarity?

Greetings from Germany
Henrik
Greetings from Germany
Henrik