Signal Generators
Frequency error when generating/measuring a swept sine with PXI-5412/5124
I would like to generate a swept sine wave, with frequencies ranging from 1kHz to 6kHz.  The signal will be generated with a PXI-5412 Arb, fed through a bandpass filter, and measured by a PXI-5124 Scope.  I am curious to know what kind of error I can expect from both the generating and measuring devices.  If I am generating a 1kHz sine wave in the arbitrary waveform mode, can I expect this to be +/- 1Hz, 100 mHz, 10 mHz, 1 mHz?  I would also like to know the same thing for the digitizer.  I cannot find this in the specs, though it seems like it would be a function of the sampling rate error and A/D or D/A errors.  I was hoping that this would be a common question that someone else has encountered.
Mike
(6,724 Views)
Mike,
Thank you for posting on the NI Forums. The spec that you looking for is the Frequency Accuracy for the PXI-5412. The Frequency Accuracy for the ARB is spec'd at +/- 25 PPM (Parts per Million) or if you are generating a signal at 1 MHz you should expect an error of +/- 25 Hz. So for your case of 1 kHz to 6 kHz you could expect an error anywhere between +/- 25 mHz @ 1 kHz, to +/- 150 mHz @  6 kHz.
With the PXI-5124 you are looking for the Timebase Accuracy spec. This spec is listed as 25 PPM when Not Phase-Locked to the reference clock of your input device. If you do not Phase-Lock the reference clocks together you could get an error of +/- 50 mHz @ 1kHz to +/- 299 mHz. The math is not as simple as adding them together. For instance to find the worse case scenario the math would be as follows: [1 - (Accuracy Error for ARB @ Rate)] X [1 - (Accuracy Error for Scope @ Rate)] or [1-(.000150)] X [1 - (.000150)] = .9997000225 then 1 - .9997000225 = .0002999775 or approximately 300 ppm. This is the absolute worse case for the scope not being phase-locked to the reference clock of the ARB. If you do phase-lock the reference clocks together your timebase accuracy for the scope is equal to that of the reference clock (or ARB in this case), so it would be +/- 25 mHz @ 1 kHz, to +/- 150 mHz @  6 kHz.
 I have linked the detailed specifications for both the PXI-5412 and PXI-5124. For the PXI-5412 the spec for Frequency Accuracy is on page 11. The Timebase Accuracy spec for the PXI-5124 is on page 14.
NI PXI/PCI-5412 Specifications
NI PCI-5124 Specifications
NI PXI-5124 Specifications
Hope this helps, let me know if you have any other questions.
Aaron W.
CLA, CTA and CPI
(6,710 Views)
Aaron,
Thanks for the quick and thorough response.  This certainly addresses the possible timing errors for each instrument, but is there any contribution from the A/D or D/A conversions?  I'm guessing that any "noise" generated during the D/A (arb) or A/D (digitizer) conversions would likely appear as wideband or high frequency noise.  My main concern is any error in detecting the fundamental frequency in kHz - amplitudes are a distant second.  I believe your first answer should suffice, but I was hoping you could confirm my followup posting.
Mike
(6,703 Views)
Mike,
It appas as though the AC Amplitude Accuracy for the scope is spec'd at 0.7% of the signal amplitude at 50 kHz, so I would expect this to be similar at the range of 1 kHz-6 kHz. As far as the ARB goes it is spec'd at 0.006% of the amplitude range. I wouldn't expect these errors to give you much of a problem though, but it would depend on what the amplitude of your signal is going to be.
Aaron W.
CLA, CTA and CPI
(6,697 Views)