A 10% error is pretty large for a PCI card. How long have you had the card? Have you tried a self-calibration of the device within Measurement & Automation Explorer? - this may help solve the apparent error on the card itself. If the self calibration fails you may wish to consider having the card professionally calibrated or, if the worst comes to be, it may need to be repaired.
Our cards are normally shipped with a calibration period of 12-24 months. Outside of this time it cannot be garaunteed to be behaving itself 100%. And as I said before, I wouldn't consider this to be normal behaviour for this device.
If you have any other queries regarding this issue just ask.
Let me know how you get on.
Many thanks,
Rob
12-07-2004 05:18 AM
12-07-2004 09:07 AM
03-24-2006 07:54 AM
03-27-2006 09:50 AM
04-12-2006 09:39 AM
Ensure that your load is 50 ohms. If not, you'll have to adjust. When you ask for 5V, NI-FGEN assumes you have a 50 ohm load. If you have a high impedance load, you will get almost 10 Volts. If you know your output impedance you can compensate.
For newer signal generators (5421, 22, 12, 41), you can set the Load Impedance attribute to inform NI-FGEN of your load impedance and NI-FGEN will do the calculation for you.