02-25-2008 04:28 PM
02-27-2008 09:31 AM
Hello TF,
It is not possible to output three counters onto one RTSI line. An error should be thrown in this case. One counter output task will reserve the RTSI5 resource. If another counter task attempt to export onto the same RTSI5 line, error 89137 “Specified route cannot be satisfied, because it requires resources that are currently in use by another route” will occur.
You are correct to be using the trigger mode stepped, however this trigger will only work with one resource to provide the start trigger. Please reference the NI Signal Generators Help to view the topic titled NI 5412/5421/5422/5441/5442 Arbitrary Sequence Trigger Modes. Please note that if a trigger occurs while a waveform is being generated it will be ignored with this trigger mode.
You mentioned that you wanted to use three physically different lines to trigger three different waveforms. The NI 5412 does not support multiple trigger in this manner. Other NI Arbitrary Waveform Generators support Scripting. Scripting will allow for up to four separate triggers. The Signal Sources that currently support scripting are the NI 5421, 5422, 5441, and 5442.