06-18-2008 07:35 PM
06-19-2008 04:46 PM
Hello jtm,
If I had to guess I would say that your filters are probably set to the 4Hz bandwidth which is the factory default on the SCXI-1120. Please see page 2-12 of the SCXI-1120/D Manual to find the correct setting for an AC signal.
I hope this helps, and if not feel free to post back.
Have a great night!
06-20-2008 06:18 AM
If I set the bandwith to 10khz will that interfere with the DC readings. In other words will I have to switch back to 4khz to read a DC signal?
I hope to get back to this next week and test your suggestion. Thanks!!
06-23-2008 11:16 AM
Hello jtm,
The 4 Hz and 10 kHz filters are low pass so technically speaking the answer is no; you can read DC signals with either setting. However, the 4 Hz filter is provided to filter out noise above 4 Hz when measuring DC signals. If you don't have much noise in your system it may not be necessary to change the jumpers back to the 4 Hz filter setting, but in most cases your measurement would benefit from doing so. That being said, these filters are set on a channel by channel basis so you can always configure your channels to measure both AC and DC at the same time if that is what you were asking.
Cheers,
06-24-2008 11:24 AM