Hello JoshB,
Thanks for your post!
I was looking more into your question about only taking a single point calibration instead of a two point. Unfortunately
you will have to take a two point calibration
in Signal Express for this to work. It calibrates similar to how you would make a custom scale. It needs at least two different points to make this calibration so that there is a slope to calibrate to. So what you could do instead of calibrate is simply view the error in the calibration wizard on step 3 where it tells you the difference and then programmatically
compensate for the error. There is a formula function in signal express that you can use to calculate this offset and program the difference. You mention that it does not need to be super-precise so perhaps a software compensation will work for your application. If you have LabVIEW you can also do some calibration using VI's. see the following:
Please post back if you have any other questions.
Cheers!
Corby_B
http://www.ni.com/support