LabVIEW 7 and NI Switch and DMM function, Won't compile

After installing labview 7, I get broken arrow for PXI Ni Swich. Error occurs in NiSwitch Initialize.vi when initialzing Switch. The error message indicates: "Call Library Function Node: Function not found in library"
0 Kudos
Message 1 of 3
(6,812 Views)
After installing LV 7.0 you should install NIDAQ 7.0. NISWITCH 2.0 and NIDMM 2.1.2. If you still have problems, please call our support line so our application engineers can help further.
Regards,
0 Kudos
Message 2 of 3
(6,811 Views)
Thanks for you input. I did installation in the sequence you mentioned but it didn't help. With a help of apps at NI and being on the phone for soooo long, he found out that two dll files called ivi_Switch.dll and ivi_switch_32.dll were listed in VXIPNP\WINNT\bin folder as well as C:\programfiles\ivi\bin. By removing those dll from VXIPNP folder fixed the problem.

Bottom line is - leave the dll fils in C:\programfiles\ivi\bin as it is!
0 Kudos
Message 3 of 3