Switch Hardware and Software
PXI-2548 relay module calibration
#1)  Where would I find the date when my PXI-2548 relay module was first calibrated?
#2)  Are there different ways to calibrate the this relay module and if so what are they?
#3)  How often should this relay module be calibrated?
Julian
(6,844 Views)
Accepted by mrjules44
Hi Julian,
The NI PXI-2548 does not require calibration because it does not have any electronic components that can get calibrated.  Therefore, there is no calibration done, there are no ways to calibrate the module, and there is no calibration schedule.
Note that you can compensate for the insertion loss of the switch on a per path basis.  These compensations typically include the loss through cables and are performed when the board is placed in the end system.  This is more of an offset compensation for the system "switches+connectivity" rather than a board calibration; it is system dependent.  I recommend making this compensation measurement before proceeding with any automated testing.
Hope this helps!
Chad Erickson
Switches Product Support Engineer
NI - USA
Message Edited by Chad PSE on 07-30-2009 05:21 PM
(6,838 Views)