09-17-2007 12:14 PM - edited 09-17-2007 12:14 PM
Message Edited by ERT_LSU on 09-17-2007 12:16 PM
09-17-2007 01:04 PM
Hi ERT_LSU,
I believe the person you previously had contact with at NI is correct. You need a voltage reference. Without it, the DMM + signal is connected to your electrode, but the - signal is floating. Tie 1_WIRE_LO_REF to a system GND (or common reference potential that does not change) and see what happens.
Hope this helps!
Chad Erickson
Switch Product Support Engineer
NI - USA
09-17-2007 01:15 PM
Thanks for helping.
Can you just explain what is a system GND?
Should it work if I connect the 1_WIRE_LO_REF to a battery (say 5 V) and keep my power source voltage at 12 V?
I actually want to use A side as + and B side as -. But I think in that way too we can't have 24X24 values for one instance. The only way to measure all the values in one instance is possible by takig potentials for all the elctrodes and subtracting them according to the need.
As I said, voltages should be like V1-25 etc. So, the way you are suggesting is actually getting the potentials for each electrodes and getting V1-25 as V25 - V1. Please correct me if I'm wrong and you meant something else.
09-17-2007 01:51 PM
Hi ERT_LSU,
I think I understand what you are attempting to accomplish with your application. You continue to describe it as a 24x24 matrix. However, you do not need the functionality of a full matrix (which is to connect x rows to y columns simultaneously). Rather, you simply need two 24x1 multiplexers. One to select an A electrode, and another to select a B electrode. A is always routed to DMM+, and B to DMM-. As you've correctly determined, the SCXI-1127 does not have a "dual" topology like some other NI switches do, so you're using an alternative method of measuring one potential and subtracting it from another.
To answer your questions:
A system GND is a ground reference common to the entire system.
The 5 V battery may not work as a common reference potential. What's keeping the battery reference fixed relative to the power source's reference?
Can you use the power source's GND as the 1_WIRE_LO_REF?
Chad Erickson
Switch Product Support Engineer
NI - USA
09-17-2007 02:08 PM
09-17-2007 03:56 PM - edited 09-17-2007 03:56 PM
Message Edited by ERT_LSU on 09-17-2007 04:00 PM
09-19-2007
06:48 AM
- last edited on
04-29-2024
06:06 PM
by
Content Cleaner
Hi ERT_LSU,
As Chad mentioned you would be able to do this measurement on the same card with some of our other switch card options. Cards such as the SCXI 1130 and the PXI 2530 offer Dual Multiplexer modes that would allow you to connect both to one card. Otherwise you will need two SCXI 1127 modules.