05-02-2023 09:46 AM
Hi,
I've stumbled across an issue with two NI-9411s being used in a NI-9149 c-RIO to measure linear motion. Each 9411 is used one at a time and they are reporting encoder output differently. Here's the scenario:
1 - cRio chassis is being used
1 - quadrature encoder
1 - LabVIEW program
1 - DUT.
When I have 9411 - "A" plugged into the c-RIO and connected to encoder, the test item travel is reported as increasing.
When I remove 9411 - "A" and plug 9411-"B" into the c-RIO, connect the exact same encoder to it, and move the DUT through the same travel motion, the travel is reported as decreasing.
The c-RIO and 9411's are brand new.
Is there a setting on the card that could have been changed somewhere, a bad card, or something else possibly?
Any insight would be helpful.
Thanks!
05-02-2023 10:10 PM
Hi Dakota,
Sounds like an interesting problem. Are you placing 9411-A and 9411-B in exactly the same slot of the 9149 chassis? If so, given that everything else about the system is the same (ex: same LabVIEW code, same PC, etc.), then the results from the 2 modules should be the same. Also be sure to check the wiring for both of your modules is exactly the same, as the A and B phases may be swapped for the 2nd module, causing the reverse in direction.
BR
John
Add motion to LabVIEW in 30min or less - TENET EMotion
Finding it hard to source NI hardware? Try NI Trading Post
05-05-2023 04:14 PM
John,
Yes, everything about the test system is exactly the same with the exception of the NI 9411 module. I was utilizing one test system and just swapping the 9411s out. Ended up sending the module back to NI for evaluation.
Thanks!