09-12-2005 12:16 PM
09-13-2005
02:18 PM
- last edited on
07-25-2025
02:26 PM
by
Content Cleaner
Good afternoon xcomas,
The 4551 samples at a maximum rate of 204.8 kS/s. So the maximum frequency of a signal that it can measure would be theorectically half this (following Nyquist) at 102.4 kHz.
I am not certain what you mean by swept sine mode, the 4551 is a Dynamic Signal Acquisiton card, it acquires signals, you can certainly choose to send a swept sine to it, in which case the sweep can go as high at 102.4 kHz.
Have a look at the NI 4551/4552 User Manual, it should have all the specs you're looking for:
https://www.ni.com/docs/en-US/bundle/321934a/resource/321934a.pdf
Please let me know if this answers your question and if you have any further questions.
09-14-2005
08:52 AM
- last edited on
07-25-2025
02:27 PM
by
Content Cleaner
09-15-2005 10:44 AM
Dear Jack,
thanks a lot for your message. You're absolutely right with your assumptions. That is a big help. I was trying to figure out if there was any way to get into a high frequency range (MHz scale) with my instrument. I've been using it before to take low frequency measurements on soil samples (to determine electrical behaviour and properties). I guess I would need something different to measure on the MHz scale (e.g. impedance analyzer)...
thanks again for your reply...
Xavier
09-15-2005 10:46 AM
thanks a lot Minh,
see reply to Jack for further explanation on what I was trying to do, but thanks again for your time and response...
Xavier