The [c]FP-TC-120 module supports 4 voltage input ranges; +/-25 mV, +/-50 mV, +/- 100 mV and -20 to 80 mV. If your sensor provides an output voltage that is within either of these 4 ranges, then you can use the [c]FP-TC-120 module to measure it. The primary restriction is the relatively narrow maximum range of the voltage inputs (+/- 100 mV). Another thing to consider is the input impedance (see the [c]FP-TC-120 Operating Instructions for more information on how the sensor impedance in conunction with the Open Thermocouple detection feature affects the measured voltage). Finally, the [c]FP-TC-120 module should not be used with high common mode voltage between channels.
Regards,
Aaron
LabVIEW Champion, CLA, CPI