Channel Ghosting issue

Solved!
Go to solution

I have a setup and I am getting channel Ghosting. Input on channel 1 24V, Channels 2,3,4,5 all showing 12-13Volts. Is there a way to fix this in Flexlogger I have seen a fix for LabView. 

 

0 Kudos
Message 1 of 6
(604 Views)

What hardware are you using? How is it wired? How do you have the channels configured?


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 2 of 6
(519 Views)

Hi the setup is as follows. 

Using a cDAQ-9174(4-slot USB)

I have the following cards in. 

NI 9217 2 channel of RTD. No issue with this. 

NI 9209(DSUB) wired in to measure current over a 1 Ohm resistor. No issue with this. 

NI 9221 (DSUB) Have 5 channels all measuring 24V. Channel 0 to 4 allocated. If there is 24V on Channel 0 all the others are reading 12/13 Volts. 

I have however put a resistor across the inputs and this has removed the ghosting. That would indicate that the NI module is not a fan of floating channels when the measuring value is high, which was said in another thread. The setup will not normally have nothing in them when the test is running. 

I have read up that this can be sorted in LabVIEW by adjusting the scanning timing, I was wondering if that was the same for the FlexLogger. 

I am very new to NI and I have tried LabVIEW but its over complicated for doing simple Logging data jobs. I liked FlexLogger for its easy to use and off you go working. 

0 Kudos
Message 3 of 6
(507 Views)

I am less hopeful that there is a simple software way to eliminate ghosting in FlexLogger. Instead, I would call it a best practice on hardware wiring to short unused inputs to GND with a moderate-sized resistor.

 

NI needs to add a FlexLogger feature that allows users to remove ghosting by adopting the repeated channel method or adjusting the ADC conversion time and settling delay.


https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P84RSAS&l=en-US

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
Message 4 of 6
(497 Views)

Hi Cheers for the info. Its working now so will see how I get one with the test. 

0 Kudos
Message 5 of 6
(491 Views)
Solution
Accepted by Hansolo080973

@Hansolo080973 wrote:

I have however put a resistor across the inputs and this has removed the ghosting. That would indicate that the NI module is not a fan of floating channels when the measuring value is high, which was said in another thread. The setup will not normally have nothing in them when the test is running. 


Nope. ADCs do not like floating inputs. You need your voltages to have some reference to ground. I highly recommend giving this article a good read: Field Wiring and Noise Considerations for Analog Signals 