 ajain
			ajain
			01-07-2010 12:50 AM
Hello
I am working on a communications ATE and need a high speed digitizer to handle my incoming signals. I need a digitizer that has very high amplitude resolution (vertical). I have looked at the PXIe 5622 and this seems to fit my requirements quite well.
However, looking at the datasheet I am a little surprised as the specified performance varies from the rated performance. Anyone else has experienced the same? I tried to use a calculator from Analog Devices and found the effective resolution to be quite low. Can someone tell me what is the expected resolution with this board? Has someone used this board and tried to make effective bits resolution measurements?
Looking forward to a response.
Regards
Asish Jain
 Henrik_Volkers
			Henrik_Volkers
			01-07-2010 04:43 AM
What's the problem with the spec? NI anaouces XX Bit and YY Sample/S....
and in the spec you find AA Bit AT BB Sample/S ... the spec is what you can claim. However the spec might not tell you wan't , sometimes because it's not specified, moslty because it's burried in other numbers/graphs you don't expect.
 Dan_K
			Dan_K
			01-08-2010 05:25 PM
Dan King
 HSD
			HSD
			01-08-2010 06:25 PM - edited 01-08-2010 06:25 PM
The 5622 is optimized for frequency domain applications (like communications). The frontend bandwidth is between 3 MHz - 250 MHz with a 150 MS/s ADC. The resolution is 16 bits, but the effective number of bits will be based on your input signal:
ENOB = (SINAD - 1.76) / 6.02
SINAD is the ratio of the fundamental to sum of all the distortion and noise products.
I would refer to the single tone and two tone signal graphs in the specs of the 5622 to see the preformance of the frontend.
The 5622 also has a DDC (digital down converter) built into the firmware. If your signal is of a narrower bandwidth, you can use the DDC to increase the ENOB of the acquisition. The ENOB inprovement is proportional to the amount of decimation (as you decimate you reduce the bandwidth you can capture around the frequency of interest). The additional bits gained is 10 log(decimation factor) / 6.02. However, the total number of bits gained is limited to 16 bits because the DDC only operates with 16 data. For instanace if you decimate by 10 (sample at 15 MS/s), your bandwidth would be limited to 6MHz (centered at whatever frequency you program); however, you would gain 1.66 bits of ENOB.
Keep in mind that various communication applications are suppectable to other factors: Distortion, Phase Noise, and Linearity of the ADC. For many applicatons these other factors are just as important as the ENOB. The 5622 was designed with all these factors in mind.