High-Speed Digitizers

cancel
Showing results for 
Search instead for 
Did you mean: 

Does the 5112 Scope card have an onboard readable temperature device?

Scope spec suggests performing a self cal if ambient temperature changes by more than 2 degrees. I assume a Temperature device is located on card to allow determination of this?
0 Kudos
Message 1 of 2
(5,858 Views)
The 5112 does not have a readable temperature device. The specification recommends self-calibrating if the ambient temperature changes by more than 2 degrees C because within +/- 2 degrees C of the self-calibrated temperature is where we guarantee our specs. Once the board drifts beyond 2 degrees C from the self-calibrated temperature, gain and offset drift can affect your measurements. Please let me know if you have further questions.
Josh
0 Kudos
Message 2 of 2
(5,851 Views)