Connect hamamatsu c9525 to labview via USB port/communication error
thank you!
did that: *idn\r
read operation
return count 4 bytes
BCBC
before, it was 2 bytes
BC
the error still stands: timeout expired before operation completed
Message 21 of 62
(3,657 Views)
Did you verify if that command is in the manual? What about the driver you downloaded? Also, the termination character had to match what the instrument is sending. Did you change It to the correct one?
Message 22 of 62
(3,652 Views)
the read device setting command is
R<CR>
the response can be
13bytes of data if the read is successful or
BC if failure.
Where can I enter these commands; for example, if I wanted to set the LCD off,
the command is: B<0h>CR
Where may I enter this command in order for it to be sent to the source?
the "select or enter command" drop down menu : how do I use it?
what does it do?
Message 23 of 62
(3,649 Views)
What menu are you talking about? In MAX?
Message 24 of 62
(3,646 Views)
yes,in max
input/output, top right
below are
write query read clear
buttons
Message 25 of 62
(3,643 Views)
You have a text box on the write tab so enter your command there. I thought you were already doing this when you got the timeout error.
Have you tried the driver you said that you had?
Message 26 of 62
(3,639 Views)
rigt now I am at home and I cannot connect the source.
On the test panel I have 4 large buttons:
configuration, input/output,advanced, NI i/o trace
on configuration submenu:
serial settings,flow control settings, i/o settings/view attributes.
on i/o settings:
write,query,read, clear
Where, on those menus, could I find a textbox that accepts commands(apart from the small window on the top-left with *idn\r  )?
 Regarding the driver:
-the one for the operating system is working properly;
-when I searched using Instrument Driver Finder I found an example called Hamamatsu C9525.
I opened it and tried to run it but it resulted in an error which was expected given the Max error.
Now I don't have with me the power source this weekend but I took the laptop with me home because I would really like to learn how to use NIMAX and Labview for control and acquisition, so I will try to install a webcam using MAX and control it in LabView.
Hopefully, by Monday, I will be more literate in Labview.
Message 27 of 62
(3,635 Views)
That little window is what you use. Delete the default string that is there since your instrument does not support it.
A camera will be completely different than the power supply.
Message 28 of 62
(3,631 Views)
Alright:
So I'll delete the column of instructions and insert the ones which the HV source understands.
I can't wait for Monday to try this...
until then, I will try what I said: become familiar with LabView in controlling an actual device(as opposed to VI simulations).
Dennis, your patience and the quality of your advice have been outstanding!
Thank you very much for that !
Message 29 of 62
(3,623 Views)
Hello again!
in the manual, to set the high voltage output,
command: V<value1><value2><CR>
argument value 1: 0h~14h
               value2: 0h-63h
response: success: VA
             failure:BC,BA
Notes:
Input value1 and value2 in binary.
The range of value1 is from 0 to 20.
The range of value2 is from 0 to 99.
High voltage output = <value1> x 100 + <value2>
High voltage output is limited to maximum voltage or less.
As shown below, the command line can be also described in binary.
<56h><value1><value2><0Dh> ※hexadecimal representation
For example, in the case of Visual Basic, when you want to set the
high voltage output at 1000V, you can write as follows;
Dim Cmd
Cmd = “V” + Chr(10) + Chr(0) + Chr(13)
or
Dim Cmd(3) As Byte
Cmd(0) = 86 : Cmd(1) =10 : Cmd(2) = 0 : Cmd(3) = 13
I copyed and pasted
then
then
Dim Cmd
then
Dim Cmd(3) As Byte
into the selector enter command window, pressed write , pressed query
but the voltage setting remained unchanged at 0Volts
and I received:
1:write operation(cmd=\x93v\x94 + chr(10) + chr(0) + chr(13) )
return count: 38 bytes
2: read operation
return count 2 bytes
BC
how should I write this simple instruction in order to set the voltage?
what am I doing wrong?
Message 30 of 62
(3,605 Views)