Hi Steve,
It is recommended to use ibdev (as opposed to ibfind). What is the link to the post that discussed using this command too often? Going to compare the NI-Spy captures is a great step. Have you been able to take a look at what the last commands were to be sent before the lock-up occurs by pulling up the NI-SPY file after the lock-up?
Also, if you run just the C-portion of the program or just the LabVIEW portion, do the lock-ups stop occuring? This may help us to determine whether the differences in the NI-SPY capture between the different programs is affecting this behavior. And, although the lock-ups are not consistent, how often would you say they happen?
07-23-2007 06:59 AM
http://forums.ni.com/ni/board/message?board.id=140&message.id=24621 is the posting that referenced the interface being overloaded by ibdev calls. The lock-ups occur about once a week during heavy use of the system. I have not been able to determine if it is the Labview portion or the C portion yet, because the system is in production use, and requires both. I will try to see if I can set up a system where I can try to isolate which type it occurs with or if it is both. I just started using NISPY and have not had a lockup occur yet. My NISPY observations were just during normal use. Because of the posting I referenced, I thought it may have been a clue to my problem, since the two methods accessed the GPIB in different ways. Is there a prefered way? Should the GPIB connection to the instrument be closed after each GPIB command, or left open for future commands, or does it matter? Thank you for your response.
Steve
07-24-2007 12:22 PM
Hello Steve.
The recommended way is to usually close the session when you are done using the instrument. If you open and close a session everytime you want to perform a task, the program will be much slower due to the overhead in opening and closing the task. This idea spans almost all of our drivers (ie DAQ, GPIB, DMMs, etc) using the open/acquire/close architecture.
However, depending on the number of calls that you are making at the same time, you may be required to close the session before opening another session. A fairly good example of this is when you are using VISA, the high level API for the 488.2 driver. A great way to troubleshoot this and find out the maximum sessions that can be opened at once can be found at the following website:
VISA Troubleshooting Wizard
http://www.ni.com/support/labview/visa/verr5.htm
This resource may or may not directly apply to the issue that you are seeing. However, it does illustrate some limitations that may be causing the issue that you are seeing. So, in order to find out more about what you are doing, how many instruments are you controlling in the LabVIEW portion of the app? How many in the C portion? Are you using the VISA API or are you using the lower level calls such as ibdev? When you say that the use is heavy in the system when the device stops responding, can you quantify this? Are you using more instruments, making more calls, increasing the frequency of the calls, etc?
Please let me know and we will continue looking into this until we find a solution!
Brian F
Applications Engineer
National Instruments